I’m a “bit” obsessive. I wanted to share my results and equipment for anyone else who is as critical of hobby grade tests. It wasn’t cheap but relatively it is.
I picked up a used Hach Drel spectrophotometer kit on ebay for $360 shipped.
http://www.hach.com/dr-2000-spectrophotometer/product?id=7640439022
http://www.hach.com/dr-2000-spectrophotometer/product-downloads?id=7640439022
The Drel kit Specs:
DR 2000 spectrophotometer (digital LCD output)
pH meter (I don’t have a need for it)
conductivity meter (I use it for a 3rd backup)
digital titrator
Soil testing kit (not included in the Drel kit, but came in the ebay auction)
Carry cases
Many outdated reagents
2 sets matched sample cells
Flasks, DO sample bottle, grad cylinders, etc, etc.
All of the extra goodies do add up quickly if purchased separately. I did replace the halogen bulb but it was done prematurely. The manuals are all available on Hach.com as well as new reagents. I picked up reagents for alk, CA/MG hardness, nitrite, nitrate, ammonia, p04, iron(just for curiosity), iodine(also does bromine and chlorine) and pH cal pillows. All told a few hundred more but you get 100 tests for most of those so price per test is cheap. There are several more test this kit can do for saltwater but I don’t feel the need to monitor for them today.
I do use API CA and ALK for my frequent tests as well as salifert for Mg. If done properly they do give good repeatable results that are quick and easy. That said, the test kits for the other parameters are a crap shoot. For all of them you must have good lab practices to get repeatable results including using the same light source to account for color shift. Even then you still estimate the color compared to a card which takes repetition to develop an eye for it. When you factor in the manufacturers stated testing variance for each kit with most hobbyist testing procedures, it leaves little wonder why many people can’t trust their results.
Tests I use for the above Hach kit:
ALK is titration but with the digital titrator it is easy
Method 8203, not much else to comment on, it easy
CA/MG hardness takes serial dilution for the DR 2000, which adds to the complexity and chance of error
I’m still debating about switching to the EDTA titrator method 8329 (reagent 24487-00) as it is easier than the serial dilution for similar amount of testing variance
Calmagite colorimetric method 8030, test 225 & 220
Nitrite is a single reagent pillow but takes 15 minutes to develop
Diazotization method 8507, test 371
Nitrate uses 2 pillows (only 50 tests per bag) and takes 7 minutes to develop
Cadmium reduction method 8171, MidRange test 353 but reagent also works for HiRange
Nitrite is a factor in the nitrate test results but typically is very small factor
Ammonia nessler or salicylate (I went nessler due to reagent cost but should have went salicylate for ease and ability to use w/ ammonia-locks)
Nessler method 8038, test 380
Po4 use one pillow and 2 minutes to develop
Ascorbic method 8048, test 490
Dead on repeatable results down to .05, any less would require I set aside specific testing equipment (cells, grad cylinder, etc) for just this test.
Iron Total uses one pillow and 3 minutes to develop
FerroVer Method 8008, test 265
Will most likely result in 0.00 results for most tanks
Iodine uses one pillow and 3 minutes to develop
DPD method 8031, test 240
Anyone who had and can remember general lab practices has the basic skills to use this kit. Things like triple rinse with ro/di, meniscus, accurate measuring and attention to detail. One note of warning. The sample sizes for most of the tests is 25ml. A great thing as it decreases the error factor but it also mean more reagent. In many cases the reagent is toxic or highly reactive. Cadmium(nitrate) alone is not safe but when you add in nessler(ammonia), the strong acids and basic chemicals you need to be careful. All or most of the above test also have standards, so you can verify your results. I’m only a bit nuts, the cost for these standards is too much. Having many other test kits and a long enough testing phase I’ve compared results from each for long enough to be very secure with the Hach results.
So it has a large upfront cost, it is not quick to use, it’s big, it’s not in an attractive green or red package, nor does it have a cult following in the hobby. What it does do is combine all of the functionality of the other electronic testers and titration methods, removing their inaccuracy providing digital results for a price per test that is equal or better. If you compare the 3 common Hanna checkers initial cost, their reagent cost per test, marginal accuracy (the entire reason for testing) to the Hach kit plus the ability to test several other parameters… you make the decision.
So I open up the debate, questions, comments, pointing out I’m nuts, or just I’m wrong.
###
#
#Part 2 - Temperature and Salinity
#http://www.chicagoreefs.com/forums/...-part-2-Temp-and-Salinity&p=242530#post242530
#
###
I picked up a used Hach Drel spectrophotometer kit on ebay for $360 shipped.
http://www.hach.com/dr-2000-spectrophotometer/product?id=7640439022
http://www.hach.com/dr-2000-spectrophotometer/product-downloads?id=7640439022
The Drel kit Specs:
DR 2000 spectrophotometer (digital LCD output)
pH meter (I don’t have a need for it)
conductivity meter (I use it for a 3rd backup)
digital titrator
Soil testing kit (not included in the Drel kit, but came in the ebay auction)
Carry cases
Many outdated reagents
2 sets matched sample cells
Flasks, DO sample bottle, grad cylinders, etc, etc.
All of the extra goodies do add up quickly if purchased separately. I did replace the halogen bulb but it was done prematurely. The manuals are all available on Hach.com as well as new reagents. I picked up reagents for alk, CA/MG hardness, nitrite, nitrate, ammonia, p04, iron(just for curiosity), iodine(also does bromine and chlorine) and pH cal pillows. All told a few hundred more but you get 100 tests for most of those so price per test is cheap. There are several more test this kit can do for saltwater but I don’t feel the need to monitor for them today.
I do use API CA and ALK for my frequent tests as well as salifert for Mg. If done properly they do give good repeatable results that are quick and easy. That said, the test kits for the other parameters are a crap shoot. For all of them you must have good lab practices to get repeatable results including using the same light source to account for color shift. Even then you still estimate the color compared to a card which takes repetition to develop an eye for it. When you factor in the manufacturers stated testing variance for each kit with most hobbyist testing procedures, it leaves little wonder why many people can’t trust their results.
Tests I use for the above Hach kit:
ALK is titration but with the digital titrator it is easy
Method 8203, not much else to comment on, it easy
CA/MG hardness takes serial dilution for the DR 2000, which adds to the complexity and chance of error
I’m still debating about switching to the EDTA titrator method 8329 (reagent 24487-00) as it is easier than the serial dilution for similar amount of testing variance
Calmagite colorimetric method 8030, test 225 & 220
Nitrite is a single reagent pillow but takes 15 minutes to develop
Diazotization method 8507, test 371
Nitrate uses 2 pillows (only 50 tests per bag) and takes 7 minutes to develop
Cadmium reduction method 8171, MidRange test 353 but reagent also works for HiRange
Nitrite is a factor in the nitrate test results but typically is very small factor
Ammonia nessler or salicylate (I went nessler due to reagent cost but should have went salicylate for ease and ability to use w/ ammonia-locks)
Nessler method 8038, test 380
Po4 use one pillow and 2 minutes to develop
Ascorbic method 8048, test 490
Dead on repeatable results down to .05, any less would require I set aside specific testing equipment (cells, grad cylinder, etc) for just this test.
Iron Total uses one pillow and 3 minutes to develop
FerroVer Method 8008, test 265
Will most likely result in 0.00 results for most tanks
Iodine uses one pillow and 3 minutes to develop
DPD method 8031, test 240
Anyone who had and can remember general lab practices has the basic skills to use this kit. Things like triple rinse with ro/di, meniscus, accurate measuring and attention to detail. One note of warning. The sample sizes for most of the tests is 25ml. A great thing as it decreases the error factor but it also mean more reagent. In many cases the reagent is toxic or highly reactive. Cadmium(nitrate) alone is not safe but when you add in nessler(ammonia), the strong acids and basic chemicals you need to be careful. All or most of the above test also have standards, so you can verify your results. I’m only a bit nuts, the cost for these standards is too much. Having many other test kits and a long enough testing phase I’ve compared results from each for long enough to be very secure with the Hach results.
So it has a large upfront cost, it is not quick to use, it’s big, it’s not in an attractive green or red package, nor does it have a cult following in the hobby. What it does do is combine all of the functionality of the other electronic testers and titration methods, removing their inaccuracy providing digital results for a price per test that is equal or better. If you compare the 3 common Hanna checkers initial cost, their reagent cost per test, marginal accuracy (the entire reason for testing) to the Hach kit plus the ability to test several other parameters… you make the decision.
So I open up the debate, questions, comments, pointing out I’m nuts, or just I’m wrong.
###
#