For last 4 years I use 90 gal. per day SpectraPure RODI. When my TDS reaches 0.01 then I'm replacing my Sediment, Carbon and DI resins. Did't replace RO membrane yet but I will this time. My question is really simple. I have always replace them with SpectraPure brand but they are quite expensive, for 4 of those cartridges I will pay around 140 $ ( best they have) Do you guys have suggestions or should I stick what works for me? I don't want to compromise health of my tank but just looking to save some money… Thanks for all advice.
To your original question - Spectrapure uses standard-sized filters, so yes, you can get refills elsewhere.
But let's back up a little...
Your TDS meter reads in ppm - you're not reading 1 one-hundredth of a part per million - so when your DI water goes above 0 or 1 ppm its time to change the resin. The TDS of your DI water tells you nothing about the need to replace your sediment filter or carbon filter, not the membrane.
Here's a cut/paste from the FAQ's on our website:
A good rule of thumb is to replace your sediment filter and carbon block after six months. A more precise way to maximize the usable life of these two filters is to use a pressure gauge to identify when pressure reaching the membrane starts to decline. This is your indication one or both of the filters is beginning to clog.
Also be cognizant of the chlorine capacity of the carbon block. A good 0.5 micron carbon block for example will remove much of the chlorine from 20,000 gallons of tap water presented at 1 gpm. Some original equipment suppliers commonly provide carbon cartridges rated at 2,000 to 6,000 gallons. Remember that all the water you process, both waste water and purified water, go through the carbon block.
Regarding your RO membrane and DI resin, use your total dissolved solids (TDS) meter to measure, record, and track the TDS (expressed in parts per million [ppm]) in three places: 1) tap water, 2) after the RO but before the DI, and 3) after the DI.
The TDS in your tap water will likely range from about 50 ppm to upwards of 1,000 ppm. Common readings are 100 to 400 ppm. So for sake of discussion, let's say your tap water reads 400 ppm. That means that for every million parts of water, you have 400 parts of dissolved solids. How do we go about getting that TDS reading down to somewhere near zero?
If you do some experimenting with your TDS meter, you'll note that your sediment filter and carbon block (collectively called “prefilters” because the treat the water before it reaches the membrane) do very little to remove dissolved solids. So with your tap water at 400 ppm, you can measure the water at the “in” port on your RO housing and you'll see it is still approximately 400 ppm.
The RO membrane is really the workhorse of the system. It removes most of the TDS, some membranes to a greater extent than others. For instance, 100 gpd Filmtec membranes have a rejection rate of 90% (i.e., they reject 90% of the dissolved solids in the feed water). So the purified water coming from your 100 gpd membrane would be about 40 ppm (a 90% reduction). Filmtec 75 gpd (and below) membranes produce purified water (a.k.a. “permeate”) more slowly, but have a higher rejection rate (96 to 98%). The lifespan of a RO membrane is dependent upon how much water you run through it, and how dirty the water is. Membranes can function well for a year, two years, or more. To test the membrane, measure the TDS in the water coming in to the membrane, and in the purified water (permeate) produced by the membrane. Compare that to the membrane’s advertised rejection rate, and to the same reading you recorded when the membrane was new. Membranes also commonly produce purified water more slowly as their function declines.
After the RO membrane, water will flow to your DI housing. DI resin in good condition will reduce the TDS in the RO water down to 0 or 1 ppm. When the DI output starts creeping up from 0 or 1 ppm, you know that your resin needs to be replaced. Sometimes people complain that their DI resin didn't last very long. Often the culprit is a malfunctioning RO membrane sending the DI resin “dirty” water. This will exhaust the resin quicker than would otherwise have been the case. Sometimes the problem is poor quality resin – remember that all resins are not created equal.
Additionally, don’t forget to sanitize the entire system at least once per year, and wash and lube your housing o-rings with food-grade silicone grease every filter change.
Russ